CVP is being displayed as a mean on most monitors. There is no "correction" for this with regards to inspiration on the Marquette's or Hewlett Packard/Agilent product.. other than the averaging function associated with obtaining the mean.
With regards to PA pressures, no.. the monitor (again, HP and Marquette) doesnt automatically distingish between inspiration and expiration in displaying the values. How do you know? Well, you dont have a mechanism to tell the monitor that the patient is recieving positive pressure ventilation, which directly impacts how you read PCWP. Remember, end expiration on a non-ventilated patient is "highest" and on a positive pressure ventilated patient it is the "lowest" value on the PA tracing.
So.. No.. the machine doesnt correct for inspiration. In the vast majority of patients there is minimal effect on PA, ART and CVP pressures by respiration (unless they are coughing), so just chart the value at any given moment.
Dave
terrie said:
I understand that PCWP measurements need to be taken at the end of expiration. Does this also apply to taking PAD and CVP pressures?
Are monitors that display pulmonary artery systolic and diastolic pressures able to automatically display these values at end of expiration?
CVP is being displayed as a mean on most monitors. There is no "correction" for this with regards to inspiration on the Marquette's or Hewlett Packard/Agilent product.. other than the averaging function associated with obtaining the mean.
With regards to PA pressures, no.. the monitor (again, HP and Marquette) doesnt automatically distingish between inspiration and expiration in displaying the values. How do you know? Well, you dont have a mechanism to tell the monitor that the patient is recieving positive pressure ventilation, which directly impacts how you read PCWP. Remember, end expiration on a non-ventilated patient is "highest" and on a positive pressure ventilated patient it is the "lowest" value on the PA tracing.
So.. No.. the machine doesnt correct for inspiration. In the vast majority of patients there is minimal effect on PA, ART and CVP pressures by respiration (unless they are coughing), so just chart the value at any given moment.
Wow.. we all like being validated from time to time.
Dave
Dr. Andrew B. Chung said:
Dave S said:
CVP is being displayed as a mean on most monitors. There is no "correction" for this with regards to inspiration on the Marquette's or Hewlett Packard/Agilent product.. other than the averaging function associated with obtaining the mean.
With regards to PA pressures, no.. the monitor (again, HP and Marquette) doesnt automatically distingish between inspiration and expiration in displaying the values. How do you know? Well, you dont have a mechanism to tell the monitor that the patient is recieving positive pressure ventilation, which directly impacts how you read PCWP. Remember, end expiration on a non-ventilated patient is "highest" and on a positive pressure ventilated patient it is the "lowest" value on the PA tracing.
So.. No.. the machine doesnt correct for inspiration. In the vast majority of patients there is minimal effect on PA, ART and CVP pressures by respiration (unless they are coughing), so just chart the value at any given moment.
Dave S <[email hidden]> wrote in message news:<[email hidden]>...
Quoted message said:
CVP is being displayed as a mean on most monitors. There is no "correction" for this with regards to inspiration on the Marquette's or Hewlett Packard/Agilent product.. other than the averaging function associated with obtaining the mean.
With regards to PA pressures, no.. the monitor (again, HP and Marquette) doesnt automatically distingish between inspiration and expiration in displaying the values. How do you know? Well, you dont have a mechanism to tell the monitor that the patient is recieving positive pressure ventilation, which directly impacts how you read PCWP. Remember, end expiration on a non-ventilated patient is "highest" and on a positive pressure ventilated patient it is the "lowest" value on the PA tracing.
So.. No.. the machine doesnt correct for inspiration. In the vast majority of patients there is minimal effect on PA, ART and CVP pressures by respiration (unless they are coughing), so just chart the value at any given moment.
Dave
I thought that the effect of respiration on a non-ventilated patient would be a variation of about 10mm in pressure. Is this about right?
Isn't it necessary to verify that in a non-ventilated patient you are recording the highest pressure value in the respirtory cycle, or you numbers may be way off(recording a pcwp of 10 rather than 20)?
If you noticed, my closing paragraph omitted mention of the wedge when saying to "ignore resps".. it addressed cvp, abp and pa, NOT pcwp. So you and I are both correct, and I apologise for not being more specific in reinforcing that point.
Dave
terrie said:
Dave S <[email hidden]> wrote in message news:- <[email hidden]>...
Quoted message said:
CVP is being displayed as a mean on most monitors. There is no "correction" for this with regards to inspiration on the Marquette's or Hewlett Packard/Agilent product.. other than the averaging function associated with obtaining the mean.
With regards to PA pressures, no.. the monitor (again, HP and Marquette) doesnt automatically distingish between inspiration and expiration in displaying the values. How do you know? Well, you dont have a mechanism to tell the monitor that the patient is recieving positive pressure ventilation, which directly impacts how you read PCWP. Remember, end expiration on a non-ventilated patient is "highest" and on a positive pressure ventilated patient it is the "lowest" value on the PA tracing.
So.. No.. the machine doesnt correct for inspiration. In the vast majority of patients there is minimal effect on PA, ART and CVP pressures by respiration (unless they are coughing), so just chart the value at any given moment.
Dave
I thought that the effect of respiration on a non- ventilated patient would be a variation of about 10mm in pressure. Is this about right?
Isn't it necessary to verify that in a non-ventilated patient you are recording the highest pressure value in the respirtory cycle, or you numbers may be way off(recording a pcwp of 10 rather than 20)?
I'd still be interested to know why you think ignoring respiration is ok for measuring pa and cvp. I would think you would get the same ~10mm difference in pressure between inspiration and expiration and that would not be a minimal effect.
Terrie
Dave S <[email hidden]> wrote in message news:<_- [email hidden]>...
Quoted message said:
If you noticed, my closing paragraph omitted mention of the wedge when saying to "ignore resps".. it addressed cvp, abp and pa, NOT pcwp. So you and I are both correct, and I apologise for not being more specific in reinforcing that point.
Dave
terrie said:
Dave S <[email hidden]> wrote in message news:<[email hidden]- .net>...
Quoted message said:
CVP is being displayed as a mean on most monitors. There is no "correction" for this with regards to inspiration on the Marquette's or Hewlett Packard/Agilent product.. other than the averaging function associated with obtaining the mean.
With regards to PA pressures, no.. the monitor (again, HP and Marquette) doesnt automatically distingish between inspiration and expiration in displaying the values. How do you know? Well, you dont have a mechanism to tell the monitor that the patient is recieving positive pressure ventilation, which directly impacts how you read PCWP. Remember, end expiration on a non- ventilated patient is "highest" and on a positive pressure ventilated patient it is the "lowest" value on the PA tracing.
So.. No.. the machine doesnt correct for inspiration. In the vast majority of patients there is minimal effect on PA, ART and CVP pressures by respiration (unless they are coughing), so just chart the value at any given moment.
Dave
I thought that the effect of respiration on a non- ventilated patient would be a variation of about 10mm in pressure. Is this about right?
Isn't it necessary to verify that in a non-ventilated patient you are recording the highest pressure value in the respirtory cycle, or you numbers may be way off(recording a pcwp of 10 rather than 20)?
The other values arent impacted as much by the tidaling of intrathoracic pressures.. when they are.. that means something else is going on. Pulsus Paradoxus for instance, when you have a greater than 10mm/hg variance in your arterial pressure due to respirations, is an abnormal finding.
I agree that the CVP is on the "more variable" end of things with regards to respirations (in that its a filling pressure, like the wedge is) but I cant ever recall a nurse freezing a tracing and manually reading a CVP to the end-expiratory point like one does with a wedge pressure.
Dave
terrie said:
Hi Dave,
I'd still be interested to know why you think ignoring respiration is ok for measuring pa and cvp. I would think you would get the same ~10mm difference in pressure between inspiration and expiration and that would not be a minimal effect.
Terrie
Dave S <[email hidden]> wrote in message news:- <[email hidden]>...
Quoted message said:
If you noticed, my closing paragraph omitted mention of the wedge when saying to "ignore resps".. it addressed cvp, abp and pa, NOT pcwp. So you and I are both correct, and I apologise for not being more specific in reinforcing that point.
Dave
terrie said:
Dave S <[email hidden]> wrote in message news- :<[email hidden]>...
>CVP is being displayed as a mean on most monitors. There >is no "correction" for this with regards to inspiration >on the Marquette's or Hewlett Packard/Agilent product.. >other than the averaging function associated with >obtaining the mean. > >With regards to PA pressures, no.. the monitor (again, >HP and Marquette) doesnt automatically distingish >between inspiration and expiration in displaying the >values. How do you know? Well, you dont have a mechanism >to tell the monitor that the patient is recieving >positive pressure ventilation, which directly impacts >how you read PCWP. Remember, end expiration on a non- >ventilated patient is "highest" and on a positive >pressure ventilated patient it is the "lowest" value on >the PA tracing. > >So.. No.. the machine doesnt correct for inspiration. In >the vast majority of patients there is minimal effect on >PA, ART and CVP pressures by respiration (unless they >are coughing), so just chart the value at any given >moment. > >Dave >
I thought that the effect of respiration on a non- ventilated patient would be a variation of about 10mm in pressure. Is this about right?
Isn't it necessary to verify that in a non-ventilated patient you are recording the highest pressure value in the respirtory cycle, or you numbers may be way off(recording a pcwp of 10 rather than 20)?