  1. I understand that PCWP measurements need to be taken at the
    end of expiration. Does this also apply to taking PAD and
    CVP pressures?

    Are monitors that display pulmonary artery systolic and
    diastolic pressures able to automatically display these
    values at end of expiration?

    Thanks,

    Terrie

  2. CVP is being displayed as a mean on most monitors. There is
    no "correction" for this with regards to inspiration on the
    Marquette's or Hewlett Packard/Agilent product.. other than
    the averaging function associated with obtaining the mean.

    With regards to PA pressures, no.. the monitor (again, HP
    and Marquette) doesnt automatically distingish between
    inspiration and expiration in displaying the values. How do
    you know? Well, you dont have a mechanism to tell the
    monitor that the patient is recieving positive pressure
    ventilation, which directly impacts how you read PCWP.
    Remember, end expiration on a non-ventilated patient is
    "highest" and on a positive pressure ventilated patient it
    is the "lowest" value on the PA tracing.

    So.. No.. the machine doesnt correct for inspiration. In the
    vast majority of patients there is minimal effect on PA, ART
    and CVP pressures by respiration (unless they are coughing),
    so just chart the value at any given moment.

    Dave

    terrie said:

    I understand that PCWP measurements need to be taken at
    the end of expiration. Does this also apply to taking PAD
    and CVP pressures?

    Are monitors that display pulmonary artery systolic and
    diastolic pressures able to automatically display these
    values at end of expiration?

    Thanks,

    Terrie

  3. Dave S said:

    CVP is being displayed as a mean on most monitors. There
    is no "correction" for this with regards to inspiration
    on the Marquette's or Hewlett Packard/Agilent product..
    other than the averaging function associated with
    obtaining the mean.

    With regards to PA pressures, no.. the monitor (again, HP
    and Marquette) doesnt automatically distingish between
    inspiration and expiration in displaying the values. How
    do you know? Well, you dont have a mechanism to tell the
    monitor that the patient is recieving positive pressure
    ventilation, which directly impacts how you read PCWP.
    Remember, end expiration on a non-ventilated patient is
    "highest" and on a positive pressure ventilated patient it
    is the "lowest" value on the PA tracing.

    So.. No.. the machine doesnt correct for inspiration. In
    the vast majority of patients there is minimal effect on
    PA, ART and CVP pressures by respiration (unless they are
    coughing), so just chart the value at any given moment.

    Dave

    Would concur :-)

    Servant to the humblest person in the universe,

    Andrew

    --
    Dr. Andrew B. Chung, MD/PhD
    Board-Certified Cardiologist
    heartmdphd.comheartmdphd.com

    **
    Who is the humblest person in the universe?
    makeashorterlink.commakeashorterlink.com

    What is all this about?
    makeashorterlink.commakeashorterlink.com

    Is this spam?
    makeashorterlink.commakeashorterlink.com

  4. Wow.. we all like being validated from time to time.

    Dave

    Dr. Andrew B. Chung said:
    Dave S said:

    CVP is being displayed as a mean on most monitors. There
    is no "correction" for this with regards to inspiration
    on the Marquette's or Hewlett Packard/Agilent product..
    other than the averaging function associated with
    obtaining the mean.

    With regards to PA pressures, no.. the monitor (again, HP
    and Marquette) doesnt automatically distingish between
    inspiration and expiration in displaying the values. How
    do you know? Well, you dont have a mechanism to tell the
    monitor that the patient is recieving positive pressure
    ventilation, which directly impacts how you read PCWP.
    Remember, end expiration on a non-ventilated patient is
    "highest" and on a positive pressure ventilated patient it
    is the "lowest" value on the PA tracing.

    So.. No.. the machine doesnt correct for inspiration. In
    the vast majority of patients there is minimal effect on
    PA, ART and CVP pressures by respiration (unless they are
    coughing), so just chart the value at any given moment.

    Dave

    Would concur :-)

    Servant to the humblest person in the universe,

    Andrew

    --
    Dr. Andrew B. Chung, MD/PhD Board-Certified Cardiologist
    heartmdphd.comheartmdphd.com

    ** Who is the humblest person in the universe?
    makeashorterlink.commakeashorterlink.com

    What is all this about?
    makeashorterlink.commakeashorterlink.com

    Is this spam? makeashorterlink.commakeashorterlink.com

  5. Dave S <[email hidden]> wrote in message news:<[email hidden]>...

    Quoted message said:

    CVP is being displayed as a mean on most monitors. There
    is no "correction" for this with regards to inspiration
    on the Marquette's or Hewlett Packard/Agilent product..
    other than the averaging function associated with
    obtaining the mean.

    With regards to PA pressures, no.. the monitor (again, HP
    and Marquette) doesnt automatically distingish between
    inspiration and expiration in displaying the values. How
    do you know? Well, you dont have a mechanism to tell the
    monitor that the patient is recieving positive pressure
    ventilation, which directly impacts how you read PCWP.
    Remember, end expiration on a non-ventilated patient is
    "highest" and on a positive pressure ventilated patient it
    is the "lowest" value on the PA tracing.

    So.. No.. the machine doesnt correct for inspiration. In
    the vast majority of patients there is minimal effect on
    PA, ART and CVP pressures by respiration (unless they are
    coughing), so just chart the value at any given moment.

    Dave

    I thought that the effect of respiration on a non-ventilated
    patient would be a variation of about 10mm in pressure. Is
    this about right?

    Isn't it necessary to verify that in a non-ventilated
    patient you are recording the highest pressure value in the
    respirtory cycle, or you numbers may be way off(recording a
    pcwp of 10 rather than 20)?

    Terrie

  6. If you noticed, my closing paragraph omitted mention of
    the wedge when saying to "ignore resps".. it addressed
    cvp, abp and pa, NOT pcwp. So you and I are both correct,
    and I apologise for not being more specific in reinforcing
    that point.

    Dave

    terrie said:

    Dave S <[email hidden]> wrote in message news:-
    <[email hidden]>...

    Quoted message said:

    CVP is being displayed as a mean on most monitors. There
    is no "correction" for this with regards to inspiration
    on the Marquette's or Hewlett Packard/Agilent product..
    other than the averaging function associated with
    obtaining the mean.

    With regards to PA pressures, no.. the monitor (again, HP
    and Marquette) doesnt automatically distingish between
    inspiration and expiration in displaying the values. How
    do you know? Well, you dont have a mechanism to tell the
    monitor that the patient is recieving positive pressure
    ventilation, which directly impacts how you read PCWP.
    Remember, end expiration on a non-ventilated patient is
    "highest" and on a positive pressure ventilated patient it
    is the "lowest" value on the PA tracing.

    So.. No.. the machine doesnt correct for inspiration. In
    the vast majority of patients there is minimal effect on
    PA, ART and CVP pressures by respiration (unless they are
    coughing), so just chart the value at any given moment.

    Dave

    I thought that the effect of respiration on a non-
    ventilated patient would be a variation of about 10mm in
    pressure. Is this about right?

    Isn't it necessary to verify that in a non-ventilated
    patient you are recording the highest pressure value in
    the respirtory cycle, or you numbers may be way
    off(recording a pcwp of 10 rather than 20)?

    Terrie

  7. Hi Dave,

    I'd still be interested to know why you think ignoring
    respiration is ok for measuring pa and cvp. I would think
    you would get the same ~10mm difference in pressure
    between inspiration and expiration and that would not be a
    minimal effect.

    Terrie

    Dave S <[email hidden]> wrote in message news:<_-
    [email hidden]>...

    Quoted message said:

    If you noticed, my closing paragraph omitted mention of
    the wedge when saying to "ignore resps".. it addressed
    cvp, abp and pa, NOT pcwp. So you and I are both correct,
    and I apologise for not being more specific in reinforcing
    that point.

    Dave

    terrie said:

    Dave S <[email hidden]> wrote in message
    news:<[email hidden]-
    .net>...

    Quoted message said:

    CVP is being displayed as a mean on most monitors. There
    is no "correction" for this with regards to inspiration
    on the Marquette's or Hewlett Packard/Agilent product..
    other than the averaging function associated with
    obtaining the mean.

    With regards to PA pressures, no.. the monitor (again,
    HP and Marquette) doesnt automatically distingish
    between inspiration and expiration in displaying the
    values. How do you know? Well, you dont have a mechanism
    to tell the monitor that the patient is recieving
    positive pressure ventilation, which directly impacts
    how you read PCWP. Remember, end expiration on a non-
    ventilated patient is "highest" and on a positive
    pressure ventilated patient it is the "lowest" value on
    the PA tracing.

    So.. No.. the machine doesnt correct for inspiration. In
    the vast majority of patients there is minimal effect on
    PA, ART and CVP pressures by respiration (unless they
    are coughing), so just chart the value at any given
    moment.

    Dave

    I thought that the effect of respiration on a non-
    ventilated patient would be a variation of about 10mm in
    pressure. Is this about right?

    Isn't it necessary to verify that in a non-ventilated
    patient you are recording the highest pressure value in
    the respirtory cycle, or you numbers may be way
    off(recording a pcwp of 10 rather than 20)?

    Terrie

  8. The other values arent impacted as much by the tidaling
    of intrathoracic pressures.. when they are.. that means
    something else is going on. Pulsus Paradoxus for
    instance, when you have a greater than 10mm/hg variance
    in your arterial pressure due to respirations, is an
    abnormal finding.

    I agree that the CVP is on the "more variable" end of things
    with regards to respirations (in that its a filling
    pressure, like the wedge
    is) but I cant ever recall a nurse freezing a tracing and
    manually reading a CVP to the end-expiratory point like
    one does with a wedge pressure.

    Dave

    terrie said:

    Hi Dave,

    I'd still be interested to know why you think ignoring
    respiration is ok for measuring pa and cvp. I would think
    you would get the same ~10mm difference in pressure
    between inspiration and expiration and that would not be a
    minimal effect.

    Terrie

    Dave S <[email hidden]> wrote in message news:-
    <[email hidden]>...

    Quoted message said:

    If you noticed, my closing paragraph omitted mention of
    the wedge when saying to "ignore resps".. it addressed
    cvp, abp and pa, NOT pcwp. So you and I are both correct,
    and I apologise for not being more specific in reinforcing
    that point.

    Dave

    terrie said:

    Dave S <[email hidden]> wrote in message news-
    :<[email hidden]>...

    >CVP is being displayed as a mean on most monitors. There
    >is no "correction" for this with regards to inspiration
    >on the Marquette's or Hewlett Packard/Agilent product..
    >other than the averaging function associated with
    >obtaining the mean.
    >
    >With regards to PA pressures, no.. the monitor (again,
    >HP and Marquette) doesnt automatically distingish
    >between inspiration and expiration in displaying the
    >values. How do you know? Well, you dont have a mechanism
    >to tell the monitor that the patient is recieving
    >positive pressure ventilation, which directly impacts
    >how you read PCWP. Remember, end expiration on a non-
    >ventilated patient is "highest" and on a positive
    >pressure ventilated patient it is the "lowest" value on
    >the PA tracing.
    >
    >So.. No.. the machine doesnt correct for inspiration. In
    >the vast majority of patients there is minimal effect on
    >PA, ART and CVP pressures by respiration (unless they
    >are coughing), so just chart the value at any given
    >moment.
    >
    >Dave
    >

    I thought that the effect of respiration on a non-
    ventilated patient would be a variation of about 10mm in
    pressure. Is this about right?

    Isn't it necessary to verify that in a non-ventilated
    patient you are recording the highest pressure value in
    the respirtory cycle, or you numbers may be way
    off(recording a pcwp of 10 rather than 20)?